Install Lens
- Install the lens on the camera.
Prepare Device
- Place the unpowered device on the thermal stage with thermal pad or grease between the stage and device to improve thermal conductivity.
- Cut high insulation sheet to cover exposed thermal stage around device (1/8" polyimide sheet can be used).
- Heat the stage to 50°C and allow the device temperature to stabilize.
Focus Lens
- Adjust camera height so device is in focus.
Setup Thermalyze Software
- Enable Image Averaging and set to 64 images.
- Set the Image Sequence Record Rate to 1/sec.
Prepare a Lens Calibration Surface
- Place a 100x100 mm section of 1/8" Polyimide insulation sheet on the benchtop away from any heat source and let stabilize to ambient room temperature (corrugated box material can also be used).
Perform Uniformity Correction
- Disable Uniformity Correction.
- Open the Uniformity Correction window.
- Measure the ambient air temperature to ±1° accuracy.
- Set Low Temperature to the measured ambient temperature.
- Pick up the Lens Calibration Surface and hold by the edges so that your fingers do not heat up the top or bottom surface.
- Quickly place the surface directly under the lens (approximately 2mm away from the lens), wait for 2 seconds, and then click the Set Low Temperature Image button.
- Place the Lens Calibration Surface back on the benchtop.
- Enable Uniformity Correction
Check Uniformity Correction
- Draw a rectangle region enclosing the unpowered device.
- Note the Max temperature in the region.
- If Max is above 50°C, calculate the difference between Max and 50° (Delta = Max - 50).
- Repeat section [Perform Uniformity Correction] but increase Low Temperature by an amount equal to Delta.
Record Image Sequence
- Press the Record Images button to start recording sequence images.
- Wait 5 seconds and then power the device.
- Allow the device temperature to reach maximum steady-state temperature and then remove device power.
- Unpress the Capture Images button to stop capturing images. This will also stop recording sequence images.
- You may want to save the image sequence at this time.
Perform Emissivity Correction
- Using the sequence slider, scroll through sequence images to display the image recorded immediately before power was applied (before the device started heating up).
- Open the Emissivity Table window.
- Set Control Temperature to 50°C.
- Click Create Single 1 to create a single point emissivity table.
- Draw a rectangle region enclosing the device so that the emissivity table will be applied to pixels within the region.
- Scroll through the sequence images to view images after power was applied.