This question has been flagged
1 Reply
149 Views

The calibration files for each lens are missing from their proper File Explorer location.

Avatar
Discard
Author Best Answer

Please preform the following Camera Calibration Data Installation Procedure:

  1. Close Thermalyze.
  2. Download the camera calibration files from the link provided by Optotherm.
  3. Unzip the download using the 7-Zip File Archiver utility.
  4. Copy and Paste the FactoryCal folder to C:\Program Files\Optotherm\Thermalyze.
  5. Copy and Paste the UserCal folder to C:\Optotherm\Thermalyze.
  6. Copy and Paste the CameraSettings folder to C:\Optotherm\Thermalyze. If a CameraSettings folder is not included with your calibration data, skip this step.