A Lock-in Thermography test is opened in the Lock-in Thermography window. The overlay image shows noise in the form of vertical lines throughout the entire image.
Support Forum
This forum provides answers to commonly asked technical support questions.
If your problem is urgent and/or too complex for the forum format, email support@optotherm.com and an engineer will assist you.
Lock-in Thermography overlay image has vertical line noise
This question has been flagged
1
Reply
12
Views
Lock-in Thermography tests are performed without noise correction. Thus, the overlay of an LIT test will have noise in the image, unless noise correction is applied after the fact. Perform the following steps to remove the vertical line noise from the LIT overlay image:
- Unshow Overlay
- Turn on Noise
- Show Overlay