This question has been flagged
25 Views

When the linear stages are in "Online" control mode, clicking the arrows when the "Jog" setting is toggled does not move the stages or the stages move for a set distance and then stops before the arrow is released. However, toggling to the "[mm]" Increment setting or "stp" Step setting works. There is a jog limit set in the software. To change this, go to Test Setup > Linear Stage Control. In the Linear Stage Control window, go to Setup > Linear Stage Setup. Enter the desired limit distance under the "Jog Limit [mm]" section.